Advertisement

Broader peaks in SIM mode than in SCAN.

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

6 posts Page 1 of 1
Hello!
Recently I have started working with a GC-MS and for my surprice when I use a SIM mode the analytes tail quite much while in SCAN mode I would say they do not tail. Is this normal? Do you have any idea of the problem and how I could solve it?
Thanks in advance for your comments.
Saioa

In theory, peak shape has nothing to do with scan or sim. However, because the sensitivity by using SIM is much higher than using SCAN, you could see the tailing in SIM which you couldn't see in SCAN.

Yes, I know it has nothing to do with the mode, that is why I was very surprice to see that in SCAN (of course higher concentrations) the peak width is about 0.2-0.3 min while in SIM the width is at least 0.5 min and therefore my analytes are not well resolve. Do you think it is my own settings in the GC-MS?

Saioa,

JI2002 already explained you the likely reason for what you see... Take your SIM chromatogram and trace a line parallel to the x axis in about half of the peak width... Would the upper part look more like what you obtain with your SCAN?

Did your "peak" contain only one single compound, as you was talking about unresolved analyte peaks in SIM? If there was two (or more) analytes almost co-eluting, I guess the peak broadening in SIM was due to peak merging. You'd better check whether adequate no. of data points being acquired in each peak. Maybe you've set a relative long dwell (or scan) time in SIM mode. Just for information, a chromatographic peak should contain about 10-15 points for it to be quantitated.

Hello!
I will try to check the dwell time. Probably I have some settings in my programs which are not completely appropiate for SIM...maybe it is the dwell time as suggested. Thank you very much for the information you gave me.
Saioa
6 posts Page 1 of 1

Who is online

In total there are 228 users online :: 0 registered, 0 hidden and 228 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am

Users browsing this forum: No registered users and 228 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry