Hello !
Do anyone know how to pilot the QDa detector (waters) using Empower 3 software. the manual guide doesnt provide enough information about how you can analyse your data on empower.

Thank you !
basically it's just like using a PDA detector in Empower (if in scanning mode (3D)) or if like any other 2D detector when SIR-Mode was used.

a) make sure the MS option is checked in the project properties!
b) select the data channel you want to process
b1) if 3D-data, enter a m/z ratio to derive the chromatogram
c) create the corresponding process method (check if it is of the "MS" type")

If you also have a PDA (or other detector connected) it's also possible to just evaluate the UV trace and use the MS-Index report group to also show the MS spectra of each peak detected in the "normal" chromatogram, without processing the MS traces
2 posts Page 1 of 1

Who is online

In total there are 11 users online :: 0 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry