- Posts: 14
- Joined: Thu Jun 29, 2006 4:22 am
I have a few questions regarding alterations to the mobile phase. I use HPLC-ECD to analyse amines etc including dopamine and its metabolites, noraldrenaline, serotonin and its metabolites, two internal standards (all in all 10 compounds). I could do this fine, then I changed column to a microbore (from 4.6x150 with 5um to 1x50 with 2.5um) and I have lost all my resolution. My mobile phase was 75mM NaH2PO4, 1.7mM octane sulfonic acid, 25uM EDTA pH 3 ; 90-10 v/v acetonitrile.
I have read the following:
Increase buffer concentration = decreased retention times (RT) of all compounds
Increased Octane sulfonate concentration = increased RT of amines
Increased pH = decreased RT of metabolites
Increased organic = decrease RT of all
Increased EDTA = decreased baseline (I have otherwise also)
Are these assumptions correct. I will be beginning microdialysis soon and not only do I need resolution but my sensitivity needs to be exceptional and I don't want to blindly test 20 different Mobile Phases when I could be more refined in my approach.
Appreciate the help