Fast versus Conventional HPLC
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
4 posts
Page 1 of 1
It is quite popular to run fast HPLC using high flow rate with smaller particle size column. My question is how about the regeration of the small particle column when you have to overload the sample to get 0.1% peak quantified. I still prefer to use conventional 5 u, 4.6 mm x 15 cm column at 1 mL/min for the impurity/degradant determination. The regeneration of the column in the gradient run is fine and we can continuously run over 40 hours with consistent results. Anyone has experience of using say 3 u, short column to analyze small peaks (with overloading sample)? How long the run could last without problem. Thank you.
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
I do not see a problem that could be associated with the use of smaller columns. In our ecperience the results are equivalent.
-
- Posts: 974
- Joined: Tue Jan 11, 2005 11:37 pm
You would scale your injection volume such that the impurity peak is the same height for the fast column as for the standard column. The overload will be proportional.
A couple practical things: Reduce as much of the gradient delay volume in your system as possible when converting to a fast method. Use a detector with a fast data collection rate (scaled to the speed of the chromatogram). In fact the detector may be the rate-limiting step in your system.
A couple practical things: Reduce as much of the gradient delay volume in your system as possible when converting to a fast method. Use a detector with a fast data collection rate (scaled to the speed of the chromatogram). In fact the detector may be the rate-limiting step in your system.
Mark Tracy
Senior Chemist
Dionex Corp.
Senior Chemist
Dionex Corp.
-
- Posts: 17
- Joined: Sat Jun 25, 2005 5:51 am
I have used UPLC(Ultra fast liquid chromatography), their particle size is 1.7 micron, diameter 2.1mm. as you decrease the diameter sensitivity will increase so no need to inject more microlitre sample you shall have to inject less microlitre sample.and very low flow you can use such as 0.25ml/min because instrument has capacity to bear 15000psi pressure.as particle size is low separation is faster.so run time will be shorter.
AMITAVA GHOSH.
AMITAVA GHOSH.
4 posts
Page 1 of 1
Who is online
In total there are 10 users online :: 2 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider], Google [Bot] and 8 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider], Google [Bot] and 8 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science