Recently, I tried to estimate the limit of detection of three organophosphorus pesticide namely methamidophos, acephate and monocrotophos by carried out a series of injection (5ppb, 10 ppb, 15 ppb and 20 ppb). I found out that at the concentration of 5 ppb (standard in blank matrix), I am not able to detect any peaks and at 10 ppb, the peaks were in good shape with S/N ratio of 23.6, 14.2 and 3.7 respectively.
Can I concluded that my LOD was 10 ppb?
I did carried out recovery test at 10ppb level and the results were pretty good >90%, similarly no peaks were detected for the recovery test at 5 ppb level.
My question is:
How should I express my LOD and LOQ?
Thanks and regards.