Good morning,
I am trying to quantitate few groups of lipid molecules. Before starting I want to get suggestions from experts in this forum.
1) Can I use precursor ion scan for quantitation (For example 184 precursor ion scan for phospholipids, sphingomyelins?)
2) What is the best way for quantitation (full scan, precursor ion scan, product ion scan, MRM)? Can these scans be equally used for quantitation?
3) As I see people spike internal standard and quantitate with reference to internal standard? However I have not see people reporting about matrix effect. My question is- Can I spike internal standard to my sample and use the spiked internal standard to quantitate without knowing the matrix effect? I think some amount is lost during extraction and processing?
Please suggest me the best method for quantitation (if there is stepwise protocol for any compound that will be great)?
4) What can be the placebo matrix to measure matrix effect for human plasma sample?
5) Is the below formula the only method for quantitation?
For example:
Analyte: 40000 (Peak intensity)
Internal standard: 60000 (Peak intensity)
Conc of internal standard: 1ng/ul
Response factor = Intensity of internal standard/Conc. of internal standard
= 60,000/1
= 60,000
Conc. of analyte = Peak intensity of analyte / Response factor
= 40,000/60,000
= 0.666 ng/ul
Is this the correct way of quantitating unknown (analyte) with reference to internal standard?
Please help me by answering my queries. I seriously need to get a very reliable and precise method of quantitating unknown using internal standard.
Thank you.
Sirus