Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
My run stopped with a message ‘Lost CAN partner’ on a HPLC instrumnet.

CAn someone help me what the issue is due to and any fix for this issue to be resolved
It would be a *lot* easier if you identified the software and instrument involved.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
As far I know CAN is used only in Agilent instruments.

Take a vacuum cleaner - in powered-off instrument disconnect every CAN cable and clean away dust from sockets.

If it would not help - problem is in one of modules - try to reconfigure system and bisect - which of the module caused problem. Maybe error log exists in CS jungles (only trained technicians know where) - there have to be exact module name which did not respond.
3 posts Page 1 of 1

Who is online

In total there are 20 users online :: 1 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry