I did a lot of plasma analyses with protein precipitation. See
http://users.chartertn.net/slittle/default.htm
first topic, the text.pdf document
Initially we used a guard column very successfully getting 200 to 400 injections before changing the guard column. However, after a while we noticed that the guard columns being supplied were not as high a quality. I would test the guard column individually and found that it significantly decreased the quality of the separation. Thus, I then started to just use the column alone.
I then could get 400-800 injections before the quality of the separation decreased. I often changed (usually every 200 injections the 0.25 micron filter before the column if the backpressure increased significantly.
Often I could clean the column (10 or so column volumes) by backflushing with water/methanol (90/10) containing 1% formic followed by water/methanol (5/95) containing 1% formic, then just pure methanol. The columns would then often be OK.
A lot of our columns used in routine analysis fail due to compression of the bed. You can take the frit off the front of the column and see the compression. We have slurried some stationary phase and added it to the front of the column and brought the columns back to their initial performace spec. Sometimes if the column is very discolored at the head of the bed, we scrape out a little of the bed and then top it off with some more slurry, and carry it through the washing.
Depends on how much you value your time. Most people consider columns disposable after a reasonable number of injections. Looking at the mode of failure might give you insight into selection of a more robust column..