I can confirm that Nexera and the Prominence are the same in that you can get the purge valve to come out by simply turning it until it does so.
I can also add a story of when I experienced issues with purging a Prominence system.
I tried purging and it didn't work, I think it showed an error message something about high pressure. I took out the purge valve and had a look, couldn't see anything stuck there. Eventually I disconnected the column and tried again, a thin stream of eluent hit the interior of the column oven as I held the PEEK in that direction. So the flow when purging apparently went straight on-column, I don't think that column was used anymore as my colleague had been doing this many times during about half an hour
Eventually it all came down to the waste tubing. It was bent so that no flow could pass, which meant that it didn't matter if the purge valve was open or not, it couldn't go to waste anyway.