What is the purpose of having a glass liner in an inlet? I know it is unreactive, but does it serve any other purpose?

The reason I ask is this. I have an HP 5890 Series II I am setting up to sample gases from a plug flow reactor. The samples should be 99% nitrogen with the balance CO, CO2, O2, EtO, CH2O, and other hydrocarbons, which should not be very corrosive. I intend to use a 1/8â€

I assume your HP5890 has a Valco 1/16" 6 port valve and stainless steel tubing.

My suggestion is to connect the sample valve to the column without going through the heated inlet at all.

At the valve detach your gas line running from the valve to the inlet. In place of that line, install another line which passes directly into the oven from the valve. Then attach a 1/16" to 1/8" reducing union to the line and install your packed column into the union.

These components can be SS or FS coated SS. They are inexpensive and easy to install.

best wishes,

Rod

Thanks. I wish I had thought of that a lot sooner.

- Alek
3 posts Page 1 of 1

Who is online

In total there are 17 users online :: 2 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Ahrefs [Bot], Amazon [Bot] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry