Agilent 5975 vs. Varian 2200 GC-MS
Posted: Wed Jun 21, 2006 3:59 pm
We recently acquired an Agilent 6890 with 5975 MSD in our laboratory. We also have an Agilent 6890 with 5973 MSD. Does anyone have any comments on the stability/sensitivity/functionality of the 5975 compared to the 5973? So far we've had several issues with the 5975. The latest being the lack of baseline, negative peaks, etc. Agilent is supposed to come for yet another service call.
The decision to obtain a 5975 was very political. I advocated extensively to purchase a Varian 2200 GC-MS system as this is what I have used previously. In my experience, it was a much better system overall. More stable, less maintenance issues, significantly more sensitive, etc. Just because the sensitivity check for both systems at installation is 1 picogram of OFN, Agilent tried to argue that the 5975 is as sensitive as modern ion trap systems. I certainly do not believe this to be the case. Anyone have any comments about this?
-Aaron
The decision to obtain a 5975 was very political. I advocated extensively to purchase a Varian 2200 GC-MS system as this is what I have used previously. In my experience, it was a much better system overall. More stable, less maintenance issues, significantly more sensitive, etc. Just because the sensitivity check for both systems at installation is 1 picogram of OFN, Agilent tried to argue that the 5975 is as sensitive as modern ion trap systems. I certainly do not believe this to be the case. Anyone have any comments about this?