Response drop using capillary column of thicker film
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
4 posts
Page 1 of 1
Wondering if anyone could give me some thoughts why using thicker filmed capillary column gives a response drop of analyte when detected by MSD. I analyse a PFP derivative of drug using capillary columns of 0.25um and 1um and the latter gives an inferior response (a drop of 2 to 3 times). Will that be any ion suppression during EI as thicker column may bleed more?
-
- Posts: 5436
- Joined: Thu Oct 13, 2005 2:29 pm
A possibility -
With the thicker film the derivative is exposed to higher temperatures for longer, and might be decomposing. Does anything else change - other peaks getting bigger or smaller for instance ?
Peter
With the thicker film the derivative is exposed to higher temperatures for longer, and might be decomposing. Does anything else change - other peaks getting bigger or smaller for instance ?
Peter
Peter Apps
-
- Posts: 145
- Joined: Sat Sep 25, 2004 3:09 pm
Peter,
I tried to use shorter column with thicker film so that the PFP derivative has similar retention as in a long column with thin film. Also the reason why I use thick filmed column is because of increase in peak tailing during sample injection and thicker film seems to protect analyte from adsorption from active sites.
I tried to use shorter column with thicker film so that the PFP derivative has similar retention as in a long column with thin film. Also the reason why I use thick filmed column is because of increase in peak tailing during sample injection and thicker film seems to protect analyte from adsorption from active sites.
-
- Posts: 366
- Joined: Wed Feb 02, 2005 9:34 pm
From basic theory, a thicker film will result in more band broadening - due to stationary phase mass tranfer. Hence, you would expect shorter fatter peaks.
The shorter peaks will cause your LOD to increase - even if you are using peak area for quantitation.
Adam
The shorter peaks will cause your LOD to increase - even if you are using peak area for quantitation.
Adam
4 posts
Page 1 of 1
Who is online
In total there are 225 users online :: 1 registered, 0 hidden and 224 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: Semrush [Bot] and 224 guests
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: Semrush [Bot] and 224 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science