No peak in negative
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
11 posts
Page 1 of 1
..
-
- Posts: 103
- Joined: Sat Mar 12, 2011 4:02 pm
Try to clean or replace resistive glass capillary (transfer optics).Hi all
I can't tune my agilent 6410 LC-MS in negative mode. Due to missing low mass fragments, Autotune is not completed and I must use manual tune instead.
-
- Posts: 47
- Joined: Mon Apr 29, 2013 4:24 pm
Thanks. I replaced the capillary and cleaned other parts, but it did'nt help me.
-
- Posts: 103
- Joined: Sat Mar 12, 2011 4:02 pm
May be you have shifted mass scale in -ESI (unappropriate default MS parameters).Thanks. I replaced the capillary and cleaned other parts, but it did'nt help me.
Hypothesis: 108.7 = 112.99; 416.10 = 431.98; 579.5 = 601.98
Try to replace 301.99 to 431.98 and increase window from 16 to 48.
Also, the intensity of 118 and 322 ions is relatively low (+ESI), so try to optimize ion source parameters (decrease dry temp, etc.) and replace calibration solution.
-
- Posts: 7
- Joined: Sun Feb 26, 2017 1:49 pm
Hello,
When you switch from positive to negative mode with calibration flow on do you see signal for target negative ions decreasing rapidly?
Did you check MS2Scan?
Did you load some old autotune file?
When you switch from positive to negative mode with calibration flow on do you see signal for target negative ions decreasing rapidly?
Did you check MS2Scan?
Did you load some old autotune file?
-
- Posts: 47
- Joined: Mon Apr 29, 2013 4:24 pm
Thanks.
Actually mass parameters (i.e drying gas temperature) were not altered after last good autotune. As can be seen in the linekd pictures, the base problem is low fragment missing or their low intensity ,and that is why autotune could not be completed.
Answer to the questioned:
- Calibrant is new
- When I changed positive mode to negative, peak intensities are decreased rapidly (for standard calibrant solution).
- Last successfully passed autotune was used, but it didn't changed situation.
- I didn't use MS2Scan. How can it help?
Actually mass parameters (i.e drying gas temperature) were not altered after last good autotune. As can be seen in the linekd pictures, the base problem is low fragment missing or their low intensity ,and that is why autotune could not be completed.
Answer to the questioned:
- Calibrant is new
- When I changed positive mode to negative, peak intensities are decreased rapidly (for standard calibrant solution).
- Last successfully passed autotune was used, but it didn't changed situation.
- I didn't use MS2Scan. How can it help?
-
- Posts: 517
- Joined: Thu Jul 14, 2016 12:08 pm
- I didn't use MS2Scan. How can it help?
MS1 and MS2 scan should give the same results. If you do see the peaks in MS2 and not in MS1, the problem lies with your first quadrupole.
-
- Posts: 47
- Joined: Mon Apr 29, 2013 4:24 pm
Today I checked MS2 scan but the same results were obtained.
-
- Posts: 7
- Joined: Sun Feb 26, 2017 1:49 pm
Hello again
It may sound like a silly question but its good to check even obvious things.
ESI-L Tuning Mix (G1969-85000) and the APCI-L (G1969-85010) have the same positive ions but differ in negative. You are sure that you use the proper tuning mix.
Could you clarify the increase of peak intensities when switching polarization mode from positive to negative? You see the proper masses (neg) for few seconds and they rapidly decreases to almost zero abundance? We have such symptom year ago and cleaning of optics helps.
It may sound like a silly question but its good to check even obvious things.
ESI-L Tuning Mix (G1969-85000) and the APCI-L (G1969-85010) have the same positive ions but differ in negative. You are sure that you use the proper tuning mix.
Could you clarify the increase of peak intensities when switching polarization mode from positive to negative? You see the proper masses (neg) for few seconds and they rapidly decreases to almost zero abundance? We have such symptom year ago and cleaning of optics helps.
-
- Posts: 47
- Joined: Mon Apr 29, 2013 4:24 pm
Thanks Pepterpol
My tuning mix is correct, since about 2 years I'm using it and there was no problem about negative mode previously. My current problem (low intensity for low mass fragments in +ESI and no peak in -ESI) has occurred two weeks ago, after bad pollution of the mass spectrometer.
I apologies for my previous post, I had a mistake: When I changed positive mode to negative, peak intensities are decreased rapidly (for standard calibrant solution).
My tuning mix is correct, since about 2 years I'm using it and there was no problem about negative mode previously. My current problem (low intensity for low mass fragments in +ESI and no peak in -ESI) has occurred two weeks ago, after bad pollution of the mass spectrometer.
I apologies for my previous post, I had a mistake: When I changed positive mode to negative, peak intensities are decreased rapidly (for standard calibrant solution).
-
- Posts: 7
- Joined: Sun Feb 26, 2017 1:49 pm
Hello
If you check/clean all easily accessible parts on the calibrant way and there is no improvement i think you should dig deeply...skimmer/octopole cleaning.
There was similar problem discussed on researchgate lots of useful inforamtion:
https://www.researchgate.net/post/Is_it ... ative_mode
Good luck
If you check/clean all easily accessible parts on the calibrant way and there is no improvement i think you should dig deeply...skimmer/octopole cleaning.
There was similar problem discussed on researchgate lots of useful inforamtion:
https://www.researchgate.net/post/Is_it ... ative_mode
Good luck
11 posts
Page 1 of 1
Who is online
In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 17 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 17 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science