RI HP 1047A repair
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
2 posts
Page 1 of 1
Hi there, it needs to be calibrated or may be repaired, there are no errors coming out of it and the light intensity is 5.1 (TP1/TP2) but we are getting top flat peaks and sometimes negative peas we are running lactic acid and the MP in water, if you know someone can calibrate or fix it please reply
-
- Posts: 129
- Joined: Sat Nov 30, 2013 8:44 am
Hello.
Do you have negative peaks at the same solutions chromatograms? Negative peaks are usually OK for RI detector. Did you have no negative peaks some time ago?
Could you show some chromatograms please?
Do you have negative peaks at the same solutions chromatograms? Negative peaks are usually OK for RI detector. Did you have no negative peaks some time ago?
Could you show some chromatograms please?
Best regards,
Dmitriy A. Perlow
Dmitriy A. Perlow
2 posts
Page 1 of 1
Who is online
In total there are 11 users online :: 1 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: CF - Aimee Cichocki and 10 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: CF - Aimee Cichocki and 10 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science