I have such problems:
1. I'd like to ask, why I see a peak from methanol, when I injected the sample 20ul (methanol/water 4:1), C18, UV or DAD detector, 240nm, 1ml/min. and mobile phase also methanol/water 4:1? I injected "the same to the same" - from where this peak (h=ca 1mV, tr=4,5min.)
2. I injected a 20ul of standard fluocinolone acetonide (diluted in methanol), mobile phase methanol/water 4:1, C18, 240nm and sometimes I see separated a small peak from methanol and my peak from fluocinolone acetonide and sometimes the peak from methanol is "under" the peak of a fluocinolone acetonide. I have a question are this second situation (no separation) unacceptable and I should not make a quantitative calculation - I thing an area from methanol distorts calculation?
3. a separation methanol as sample solvent and mainly peak with water/methanol mobile phases is a typical problem and only with good condition of column the separation is OK or if this is not so important?