The monograph calls for a 250x4,6 mm column with base-deactivated end-capped C 18 silica 5um. We have used Hypersil BDS C18. The isocratic mobile phase is 22% tetrahydrofuran in water at 1 ml/min. T = 45 C. Detector set at 254 nm. The problem is poor resolution between predisolone and hydrocortisone (sample in mobile phase). The monograph requires 2.2 the best we can get is 1.77. We have made variations in the column, temperature and mobile phase.
If you have any suggestions to improve the resolution under these conditions, I would be very grateful to hear of them!