No peak in negative
Posted: Sat Mar 18, 2017 9:42 am
by alcorn
Re: No peak in negative mode in LC-MS
Posted: Tue Mar 21, 2017 2:09 pm
by sav
Hi all
I can't tune my agilent 6410 LC-MS in negative mode. Due to missing low mass fragments, Autotune is not completed and I must use manual tune instead.
Try to clean or replace resistive glass capillary (transfer optics).
Posted: Wed Mar 22, 2017 11:47 am
by alcorn
Thanks. I replaced the capillary and cleaned other parts, but it did'nt help me.
Posted: Wed Mar 22, 2017 1:10 pm
by sav
May be you have shifted mass scale in -ESI (unappropriate default MS parameters).
Hypothesis: 108.7 = 112.99; 416.10 = 431.98; 579.5 = 601.98
Try to replace 301.99 to 431.98 and increase window from 16 to 48.
Also, the intensity of 118 and 322 ions is relatively low (+ESI), so try to optimize ion source parameters (decrease dry temp, etc.) and replace calibration solution.
Posted: Wed Mar 22, 2017 2:41 pm
by pepterpol
When you switch from positive to negative mode with calibration flow on do you see signal for target negative ions decreasing rapidly?
Did you check MS2Scan?
Did you load some old autotune file?
Posted: Thu Mar 23, 2017 6:22 pm
by alcorn
Thanks.
Actually mass parameters (i.e drying gas temperature) were not altered after last good autotune. As can be seen in the linekd pictures, the base problem is low fragment missing or their low intensity ,and that is why autotune could not be completed.
Answer to the questioned:
- Calibrant is new
- When I changed positive mode to negative, peak intensities are decreased rapidly (for standard calibrant solution).
- Last successfully passed autotune was used, but it didn't changed situation.
- I didn't use MS2Scan. How can it help?
Posted: Fri Mar 24, 2017 7:48 am
by Rndirk
MS1 and MS2 scan should give the same results. If you do see the peaks in MS2 and not in MS1, the problem lies with your first quadrupole.
Posted: Fri Mar 24, 2017 10:57 am
by alcorn
Today I checked MS2 scan but the same results were obtained.
Posted: Fri Mar 24, 2017 11:34 am
by pepterpol
Hello again
It may sound like a silly question but its good to check even obvious things.
ESI-L Tuning Mix (G1969-85000) and the APCI-L (G1969-85010) have the same positive ions but differ in negative. You are sure that you use the proper tuning mix.
Could you clarify the increase of peak intensities when switching polarization mode from positive to negative? You see the proper masses (neg) for few seconds and they rapidly decreases to almost zero abundance? We have such symptom year ago and cleaning of optics helps.
Posted: Sat Mar 25, 2017 4:26 am
by alcorn
Thanks Pepterpol
My tuning mix is correct, since about 2 years I'm using it and there was no problem about negative mode previously. My current problem (low intensity for low mass fragments in +ESI and no peak in -ESI) has occurred two weeks ago, after bad pollution of the mass spectrometer.
I apologies for my previous post, I had a mistake: When I changed positive mode to negative, peak intensities are decreased rapidly (for standard calibrant solution).
Posted: Sat Mar 25, 2017 7:36 am
by pepterpol
Hello
If you check/clean all easily accessible parts on the calibrant way and there is no improvement i think you should dig deeply...skimmer/octopole cleaning.
There was similar problem discussed on researchgate lots of useful inforamtion:
https://www.researchgate.net/post/Is_it ... ative_mode
Good luck