gripper fingers failed to open/close on agilent 1100's ALS
Posted: Wed Jun 14, 2006 11:14 pm
by ntruong
Hi all,
The gripper fingers seems to die on me since it is not picking up sample vials. I called agilent and was told that I need to replace the Transport Assembly. Has anyone replaced this before? Is it easy to replace?
Thanks in advance,
Posted: Thu Jun 15, 2006 12:03 pm
by Peter Apps
Before you spend your money on a whole new assembly check all the wiring connections and that the fingers are not just jammed.
Posted: Mon Jul 17, 2006 9:32 pm
by ntruong
Thanks Peter,
I have the instrument down for a month because I didn't have time to fix it. You did save me over $1800 for part because the fingers really got jammed, not the whole transport assembly.
thanks again,
Posted: Mon Jul 17, 2006 10:19 pm
by fbreton
how did you manage to fix it?
just curious
Posted: Tue Jul 18, 2006 4:18 am
by ntruong
I removed the gripper arm, rotated the shaft to ensure that the gripper fingers opened and closed freely. Re-installed and aligned it. It seems to work fine at least for now. However, I still think the motor might be the culprit because I didn't really do anything other than "rotating the shaft" to release the lock.
Posted: Tue Jul 18, 2006 8:15 pm
by rnelson