Typically when we perform a method validaton, we evaluate the peak purity from about 210nm to 300nm. Is this enough or is it too much. What wavelength range do you use, or do you use a range around the analysis wavelength?
George Balock

we should use wavelength range 210 to 400nm
Your wavelength range for peak purity really depends upon two factors: The UV absorbance spectrum of your compound and the background absorbance of your mobile phase. If you are assaying a sunscreen with an absorbance maximum of 360 nm and you're using an acetate based mobile phase, 210-300nm won't do too well. You'd want to be between 230 - 450 nm. On the other hand, if you have an abs max at 234 with nothing beyond 250 nm and are using a phosphate-ACN MP, you might want to look at 205-270. It all depends on what you'relooking at and the conditions you're using.
Juddc,
Thanks for the input. To rephrease the question how much is enough, 50nm on either side of yoiur analysis wavelength, 100nm?
George Balock
4 posts Page 1 of 1

Who is online

In total there are 8 users online :: 0 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry