I have got a brand new column from Thermo (Syncronis C18). It turns out that I don't need to use it right away. So, I have still not opened the column. Is it okay if I just keep it at room temperature? Or I need to do something for it's storage?
-
- Posts: 326
- Joined: Wed Oct 02, 2013 12:20 pm
Hello
As you haven't opened column yet it still has storage solvent inside (you will find information about storage/shipping solvent in column certificate).
You can keep box with column (or column itself) at room temperature.
Regards
Tomasz Kubowicz
As you haven't opened column yet it still has storage solvent inside (you will find information about storage/shipping solvent in column certificate).
You can keep box with column (or column itself) at room temperature.
Regards
Tomasz Kubowicz
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
It will be fine. Or since seal is unbroken, maybe you can return it.
-
- Posts: 715
- Joined: Wed Aug 04, 2010 2:45 pm
I agree with Tomasz. It is always good to have a brand new column in the back hand, just in case.
Gerhard Kratz, Kratz_Gerhard@web.de
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Agree; always good for determining if the column is the issue when a problem shows up. I always had at least one back-up new column for each of our pharmaceutical assays, both HPLC and GC.I agree with Tomasz. It is always good to have a brand new column in the back hand, just in case.
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
Probably more true in the lab than anywhere else - when in doubt, read the manual. Even columns come with them.
Thanks,
DR

DR

-
- Posts: 41
- Joined: Mon Oct 03, 2016 10:50 am
Thank you guys.
7 posts
Page 1 of 1
Who is online
In total there are 15 users online :: 1 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 14 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 14 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science