MULTIPLE PEAKS AND BACKGROUND SUBSTRACTION CLARUS 500 GC MS
Posted: Fri Mar 17, 2017 12:23 am
Good evening everyone:
I am new to GC-MS and I would like to know how likely it is for multiple peaks to appear for a single mass. Some days ago I realized that I got two peaks for a Thymol Standard solution (just one compound). After asking Turbomass to display the chromatogram for the m/z 135 only, both peaks remained.
I've tried cleaning the inner source, and reconditioning the column, and yet, I ran a blank (just Helium) and those 2 peaks remain. I wonder if anyone has experienced this before. Also... has anyone used the Clarus 500GC background substraction utility for the MS? According to the manuals, this utility allows the user to run a blank so a background file is recorded and automatically applied to further runs when using FID or any other detector for the GC standalone mode. The manual says it is possible to calibrate the background for MS (which I already did) "WHEN YOU SUBSECUENTLY TURN THE BACKGROUND COMPENSATION ON" (Clarus 500 GC Users Guide page 6-95), but when I ran a blank just after the calibration, the two "ghost" peaks still showed up in the chromatogram.
For applying the background recorded, I pressed "Run>>Start>>ok" as usual, and when "Waiting for injection" appeared in the PC screen; then I went to the GC touchscreen and went to "Tools>>Utilities>>Background Compen", selected "Method 5" (which is the one that assumes the PC method in my case), checked "apply calibration" and pressed "start run". I wonder if this procedure was correct, since i got the very exact same chromatogram for the background recording and for the ran blank.
Thank you very much for your comments
I am new to GC-MS and I would like to know how likely it is for multiple peaks to appear for a single mass. Some days ago I realized that I got two peaks for a Thymol Standard solution (just one compound). After asking Turbomass to display the chromatogram for the m/z 135 only, both peaks remained.
I've tried cleaning the inner source, and reconditioning the column, and yet, I ran a blank (just Helium) and those 2 peaks remain. I wonder if anyone has experienced this before. Also... has anyone used the Clarus 500GC background substraction utility for the MS? According to the manuals, this utility allows the user to run a blank so a background file is recorded and automatically applied to further runs when using FID or any other detector for the GC standalone mode. The manual says it is possible to calibrate the background for MS (which I already did) "WHEN YOU SUBSECUENTLY TURN THE BACKGROUND COMPENSATION ON" (Clarus 500 GC Users Guide page 6-95), but when I ran a blank just after the calibration, the two "ghost" peaks still showed up in the chromatogram.
For applying the background recorded, I pressed "Run>>Start>>ok" as usual, and when "Waiting for injection" appeared in the PC screen; then I went to the GC touchscreen and went to "Tools>>Utilities>>Background Compen", selected "Method 5" (which is the one that assumes the PC method in my case), checked "apply calibration" and pressed "start run". I wonder if this procedure was correct, since i got the very exact same chromatogram for the background recording and for the ran blank.
Thank you very much for your comments