Posted: Tue Jun 13, 2006 5:16 am
by husss2000
how could i detect hydrocarbon presented in amine? is possible just inject the amine in the GC and i will get the result for the hydrocarbon?

info needed

Posted: Tue Jun 13, 2006 12:27 pm
by chromatographer1
what hydrocarbon? What amine?

Posted: Tue Jun 13, 2006 12:32 pm
by husss2000
we are using methyl di ethanol amine to absorbe the acidic gas from our gas and at the same time i think the amine also will absorbe the heavey hydrocarbon from our gas and i want to see how much amine could absorbed by injecting them in gc but i'm afride that amine will damge my coulmn.Many thanks.

hydrocarbon

Posted: Tue Jun 13, 2006 3:41 pm
by chromatographer1
You are looking then for C5 through about C14 at 1000ppm to a few ppm in natural gas absorbing into your amine sulfur trap components.

If you are using traditional packed column BTU analyzer then a direct injection of the amine is not recommended.

If you are using a capillary column lab GC then you could certainly inject the amine directly and determine the amount of hydrocarbon present using an internal standard.

A thick film methyl silicone or 5% Phenyl methyl silicone is recommended.

I would expect that some heavy hydrocarbons will be present but that they should be in small amounts, especially if the amine is at above ambient temperature as I would expect on a pipeline pumpling station.

best wishes,

Rod

Posted: Wed Jun 14, 2006 3:45 am
by husss2000
many thanks for your replay.I think you solve my problem.Great job by this forum hoping we continue with the same level of power .I will help alot for this forum.Many thanks.