Hornet » Mon Mar 13, 2017 4:03 pm
Depends on the application. You can't use the increased efficiency of H2 chromatographically since you can't increase the carrier flow as you would on a regular GC detector, you have to actually decrease the flow rate, or keep it equal to Helium. The vacuum gauge will read much higher pressures in the analyzer but some of that is due to the fact that most ion gauge tubes are calibrated for N2 and each gas will give a different value regardless of the actual pressure in the analyzer.
I was able to increase the sensitivity doing MRM since I can eliminate the Helium Quench gas when running Hydrogen carrier, which improves the vacuum. Tradeoffs are you have to watch to be sure you are not getting reactions with unstable analytes, but with things like PAHs and PCBs you shouldn't have to worry about it.
The new Self Cleaning Source(Jetclean) that Agilent is pushing now is simply a 0.1 to 0.3ml/minute hydrogen flow mixing with the column effluent just before it enters the analyzer, which works to scrub the source on a continual basis or as a flush at the end of the run. They even say it can be implemented on 5975s using a CI setup to introduce the H2.
The only application I have not been able to use H2 on is volatiles, since the H2 interferes with the BFB tune check by reaction which causes a high mass 96 value versus mass 95. DFTPP for semi volatile analysis I have had success with. You do have to let it system bake out for several days before the background will be low enough. The H2 scrubs out a lot of hydrocarbon and siloxane contamination that is left behind when running Helium.
Hello James, do you think i could run H2 on a 7000 for PCDD/F MRM?
I'm actually using helium and a 2x injector, 2x column. They both go into a 2 holes ferrule to source and i keep 1ml/min the main column and 0.6 ml/min the other column.
Thank you for sharing your experiences.