Quantitation/Qualifier Ions

Basic questions from students; resources for projects and reports.

4 posts Page 1 of 1
Are qualifier ions neccessary for quantitation of known compounds? I have an LC-MS method where I've seperated my two known compounds but when I look at the mass spec, I only have one m/z peak for each chromatographic peak. Is this desireable?

Since you have separated your two compounds and presumably the ions you are seeing in each case are the molecular ions for each compound then yes this is indeed a desirable scenario. By having just one ion you are maximising your signal to noise ratio.
You may not need qualifier ions for known compounds where you KNOW the compound is in the sample (e.g., PK studies). However, if you are doing contaminant analysis where you are trying to establish that your known compound is in an unknown sample, you probably need some qualifying ions.
All standard disclaimers apply: This post reflects personal opinion only and not the policies of my employer.

Thank you very much for your input!
4 posts Page 1 of 1

Who is online

In total there are 89 users online :: 0 registered, 0 hidden and 89 guests (based on users active over the past 5 minutes)
Most users ever online was 11068 on Sat Dec 06, 2025 9:55 pm

Users browsing this forum: No registered users and 89 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry