Last week I informed the group who developed this assay of the recent problems I've been having, namely that even without degassing the mobile phase, I get air bubbles. The mobile phase has dioctyl sulfosuccinate (DOSS) as an ion pairing reagent. Why this was chosen I'm not sure, but because DOSS is in the mobile phase I can not degass it. Anyway, I told the group that I was having baseline problems, and I was told to use a backpressure regulator because the DOSS will still form air bubbles in the system without it.
Today I put on the backpressure regulator and the detector's flow cell blows. The backpressure regulator (Upchurch #U-610) is rated at 40psi and the flow cell can take 145 psi. There shouldn't have been any problems.
The entire system normally runs at about 150 psi at 0.1 mL/min while running the mobile phase. However, once the backpressure regulator in put on the psi jumps to over 900. I've contacted both Upchurch and Waters. Upchurch tells me that the psi shouldn't be that high and Waters tells me not to use the backpressure regulator.
How can I correct this problem? I can't run the assay without the backpressure regulator, and I don't have time to develop a new one.
The system I'm running is as follows:
Alliance 2695 with a 2487 UV/Vis detector and a 2475 Fluorescence detector. Everything is from Waters. The 2487 is connected to the column and the 2475 is connected to the 2487, then out to waste. The backpressure regulator is connected to the waste of the 2475, and it's the 2475 that I'm using for the detection, which has the flow cell that was destroyed as well.
Who would've thought quantitating Topotecan would be so difficult.