I had been trying to build a calibration curve but was unsucessful, so I checked the RSD of my system and it was as follows:
275 - 235 - 220 - 215 - 285 (nm)
2.5 - 1.1 - 4.0 - 6.4 - 3.4 (%RSD)
This was calculated from the peak area of a single peak (no other peaks in chromatogram) across 25 samples, where there was 5 injections of 5 samples all taken from the same stock solution, total loop fill of 1ul. wavelenghts were selected from spectrum as 275 maxma, 235 minma, 220-285 flat reigons.
Long story short
I have now got the following:
275 - 235 - 220 - 215 - 285
0.6 - 0.6 - 0.5 - 0.6 - 0.6
Again across 25 samples in the same protocol.
Should I be happy with these RSD's or should I continue to try and improve them? If I should improve them what areas should look at improve the RSD?
I have a Agilent 1100 system but with a Gilson sampler and rheyodyne loop. I have already replace the the flow cell, lamps, rheyodyne and seals of injection port needle, needle.
I have lots more infomation if needed