Protien separation retention time shifts
Posted: Wed Jun 07, 2006 9:14 pm
Hi,
I have a method for measuring a 98KDa protein in buffer solutions. I was using a C8, 300A, 5um, 2.1mm x 25 cm column. Column temp = 40C. Mobile phase is a gradient from 30% to 80% B. MP A is 0.1% TFA in water, MP B is 0.1% TFA in ACN. When I initially developed this method a couple of months ago, my retention time was very solid at 16.2 min. for every injection every day. However, now, my retention time is variable and on average is getting later.
Two things had happened that I thought might have caused this problem. First, I had to switch the instrument to normal phase for a week. After I switched back and replaced the pump seals, the retention time was 17.3. Second, I started assaying some filtered whole broth samples which had a lot more proteins and salts in them. The peak of interest began shifting later until it was 20.3 min.
At this point, I tried cleaning the column according the manufacturer's recommendations and the retention times. The retention times briefly improved back to 17 min. but quickly began shifting later again. I tried a new column C8 and still saw the retention times shifting later, even with the cleaner samples of just protein in buffer.
Since I felt that the method performance had changed so much that I needed to repeat my validation, I decided I might as well try a new column. It was recommended that a C4 column might be better for a protein this large. I now am using a C4, 300A, 5um, 2.1mm x 25 cm column with all other conditions being the same. My results are the same as with the old C8 column. The peak of interest began coming out around 19 min. In two weeks it has drifted to 22 min. Any changes that I make to the gradient cause the peak to be even later - I can't seem to make it come out earlier. When I have very long runs, I can see the peak move as much as 2 minutes (e.g. from 20 to 22 minutes) during the run. This has really messed up the integration method because the peak moves out of the window.
Does any one have any ideas about why this peak keeps shifting to later times and what I can do to stabilize it?
Sorry for the very long message.
KarenJ
I have a method for measuring a 98KDa protein in buffer solutions. I was using a C8, 300A, 5um, 2.1mm x 25 cm column. Column temp = 40C. Mobile phase is a gradient from 30% to 80% B. MP A is 0.1% TFA in water, MP B is 0.1% TFA in ACN. When I initially developed this method a couple of months ago, my retention time was very solid at 16.2 min. for every injection every day. However, now, my retention time is variable and on average is getting later.
Two things had happened that I thought might have caused this problem. First, I had to switch the instrument to normal phase for a week. After I switched back and replaced the pump seals, the retention time was 17.3. Second, I started assaying some filtered whole broth samples which had a lot more proteins and salts in them. The peak of interest began shifting later until it was 20.3 min.
At this point, I tried cleaning the column according the manufacturer's recommendations and the retention times. The retention times briefly improved back to 17 min. but quickly began shifting later again. I tried a new column C8 and still saw the retention times shifting later, even with the cleaner samples of just protein in buffer.
Since I felt that the method performance had changed so much that I needed to repeat my validation, I decided I might as well try a new column. It was recommended that a C4 column might be better for a protein this large. I now am using a C4, 300A, 5um, 2.1mm x 25 cm column with all other conditions being the same. My results are the same as with the old C8 column. The peak of interest began coming out around 19 min. In two weeks it has drifted to 22 min. Any changes that I make to the gradient cause the peak to be even later - I can't seem to make it come out earlier. When I have very long runs, I can see the peak move as much as 2 minutes (e.g. from 20 to 22 minutes) during the run. This has really messed up the integration method because the peak moves out of the window.
Does any one have any ideas about why this peak keeps shifting to later times and what I can do to stabilize it?
Sorry for the very long message.
KarenJ