Wondering if anyone could give me some thoughts why using thicker filmed capillary column gives a response drop of analyte when detected by MSD. I analyse a PFP derivative of drug using capillary columns of 0.25um and 1um and the latter gives an inferior response (a drop of 2 to 3 times). Will that be any ion suppression during EI as thicker column may bleed more?

A possibility -

With the thicker film the derivative is exposed to higher temperatures for longer, and might be decomposing. Does anything else change - other peaks getting bigger or smaller for instance ?

Peter
Peter Apps

Peter,
I tried to use shorter column with thicker film so that the PFP derivative has similar retention as in a long column with thin film. Also the reason why I use thick filmed column is because of increase in peak tailing during sample injection and thicker film seems to protect analyte from adsorption from active sites.

From basic theory, a thicker film will result in more band broadening - due to stationary phase mass tranfer. Hence, you would expect shorter fatter peaks.

The shorter peaks will cause your LOD to increase - even if you are using peak area for quantitation.

Adam
4 posts Page 1 of 1

Who is online

In total there are 222 users online :: 0 registered, 0 hidden and 222 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am

Users browsing this forum: No registered users and 222 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry