Injector program validation

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi all,

To cut a long story short I've had issues with sample stability. After trying many avenues I find the only way to get around these issues is to mix the sample pre-injection using an injector program (I'm using an 1100).

Having never validated a method with an injector program I'm wondering if I need to do anything over an above a typical validation for an LC method.

My initial feeling is that if it passes the usual validation then by definition it means the injector program is validated along with it.

Any guidance greatly appreciated,

Serp

I would be inclined to have it mix portions of 0.3% acetone w/ water (from 0-100% of the Acetone solution) and see if the intermediate peak responses are the appropriate proportions of the 100% acetone solution response.

Fixed injection volume throughout, very short run time... you could do it w/o a column running water at a fixed flow rate. It would be pretty much like checking pump mixing.
Thanks,
DR
Image
2 posts Page 1 of 1

Who is online

In total there are 15 users online :: 1 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry