Figure 1: Good Blank by LC-UV
Figure 2: Good Standard by LC-UV
Figure 3: Good Blank by LC-ICPMS
Figure 4: Good Standard by ICPMS
Everything works perfectly when the waste from the UV really does go to waste. As soon as I connect the waste from the UV to the ICPMS, I get this frequent peak splitting effect:
Figure 5: Bad Blank by UV
Figure 6: Bad Standard by ICPMS
This peak splitting doesn't happen with every sample, but is random. If I run 10 replicates from the same vial, 5 might be split peaks, 5 might be okay. Also, if I bypass the UV and connect the column to the ICPMS, I get the same effect (but with no UV trace for comparison!).
The fact that these split peaks are not consistently observed in the same sample suggests it's not really in the sample, and the fact that split peaks are observed in both the analyte and the peaks observed in blanks suggests that it is not an ICPMS nebulisation/sample intake issue. Plus, the fact that I am getting split peaks in the UV as well as the ICPMS suggests that it is a physical effect occuring between the autosampler and the UV detector that is a direct result of connecting the flow to the ICPMS.
We have a number of other LC-ICPMS methods but I have never encountered this before!
Any comments?