Hello,
I have a method to analyze Phenylepthrine using HPLC based of the method here:
http://www.phenomenex.com/Application/Detail/3246
The only difference is I'm using a slightly lower flow rate at 0.7mL/min
It works wonderfully for Phenylephrine by itself,but now I am trying to analyze a solution with Phenelyephrine HCl, sodium chloride, sodium bisulfite, and edetate disodium and I have two peaks that are very close to each other.
Any tips on how to get these two peaks to separate better without drastically changing the method? I've already tried altering the flow rate and temperature...
Thanks!
Use a 5 cm Luna column. Eliminate Octanesulfonic Acid (its obsolete) and replace it with 0.05 M Phosphate buffer pH 3.0. Go gradient to tease these peaks apart.
Its apparent this Phenomenex method does not work for your sample matrix.
Thanks for the reply! I have replaced the Octanesulfonic Acid with acetic acid (pH 3.0) do you think the buffer is necessary to desperate or will the gradient surfice?
Gradient. We could only separate Acetaminophen, Phenylephrine, Doxylamine HBr, Diphenhydramine, and Chlorpheniramine HCl using gradient and wavelength shifting.
Would you mind sharing a method?
6 posts Page 1 of 1

Who is online

In total there are 33 users online :: 2 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry