Initially we observed that pump head B would suddenly drop in flow rate for no apparent reason. After trying all the obvious changes without eliminating the problem, we observed solid particles in the pump head B inlet tubing ! This observation, combined with the fact that the 1st run usually works fine and the unit fails in the 2nd run suggest that solvent A is leaking into pump head B and precipitating in the inlet line/check valve/pump head. I have been using HPLC for more than 10 years, including all major brands except for Gilson, and I have never heard of such a thing. What's more, the problem is not limited to our pump. Gilson loaned us a different 321 pump, which experienced the same problem! Now Gilson is claiming that we are simply asking too much from the pump.
I was wondering if anyone else has ever observed this in a Gilson pump, and if so, were you able to solve the problem? Is the Gilson really so limited? Isn't the whole idea of a check valve that liquid only flows in one direction? We have 3 Varian systems that run the same buffer system without any problems. Any ideas?
Byron in Madison