Hey I am working on a cannabis method for HPLC using Labsolutions.
I have created a calibration curve from 1 - 200 µg/mL and run it in triplicate. The results are linear, but then I find if I run the curve again as unknown samples the %recovery on the lower end is way out.

However, if I use exponential as the line of best fit, the %error drops for all points and the recovery is much more acceptable.

When modeling calibration curves and looking at the %error plots, how do you decide if there is a sample prep issue or systematic issue that is it ok to use different modelling? My error plots look like a very broad peak (high end has low error, then going towards middle increases and then drops to very low -% error at the low concentration end). I assume this means I am losing analyte as I go down, compounded with greater variation on smaller concentrations?

BTW I perform serial dilutions using micropipettes up to 300 µL each due to volume and cost of standards.

Thanks
Just to add some more info; for my lowest calibration point, the peak area is around 9000 and yet the curve built from that injection has a y-int of 19000. So you can see my problem at the low end. We can get negative values.

However this is completely corrected with exponential fitting.

I haven't even gotten to accuracy testing matrix issues yet... :(
What you're seeing is not unusual in a wide-range assay; ordinary least squares assumes that absolute errors are more or less constant across the entire range, whereas chromatographic methods typically have *relative* errors constant (so absolute errors are greater at the high end of the range). The problem can be sidestepped in a number of ways such as weighted least squares, transform to log-log, or polynomial fit. Check out Sections 8-10 of the Coleman & Vanetta "Statistics in Analytical Chemistry" series for more details. You can find links to the pdfs here:
- http://www.americanlaboratory.com/1403- ... lpi_4346=3
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
3 posts Page 1 of 1

Who is online

In total there are 14 users online :: 1 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry