I have a problem with my Agilent 5977E MSD.
When i pump down its waiting and after give error "VacSysFail". After that take shutdown MS.
ms menu says that:
MSD Fault Status:
Dec 384 hex 180
Advertisement
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
Hello
Can you hear turbo pump running? (or see if it starts at all with %?)
Check if analyser is closed (perhaps it is still open after column changing/maintenance).
Regards
Tomasz Kubowicz
I've been pulling my hair out because i couldn't get the same model to start pumping just a couple of hours ago! First time this happens. My fingers still hurt from pushing the side i'm not even kiddingIf your interface had the white ceramic tip inside the analyzer, the spring on that can push the door open, or prevent you from closing it completely. I had this happen on my 7000QQQ once. Took me a while to figure out what was happening
Well i managed to get it pumping today.Also before closing the side plate, push on the ceramic cone to make sure it isn't binding. If the spring inside gets a little off it can cause it to bind instead of sliding easily as it should.