Hi
The best application for methanizer to keep it work stable, linear, etc., for years and i realy mean for years, is to use back flush valve between the column and the methanizer and to avoid any hydrocarbons heavier than C1 enter to methanizer. we use Porapack column, the ellution order is CO,CH4, CO2 and then backflush. we use 4 methanizers, and one of them we operate for more than 20 years, 10-12samples in 24 hours a day, every day year around. some samples contain hydrocarbons up to C7 and higher, some contain H2S & RSH, and other impurities. the system works.
Good luck