If you can get the column out of the oven and have a look at the first two meters it will be better - you cannot bake away septum particles. With your two-column setup though that might be tricky.
Peter
Hi Peter,
I have whipped the column out this morning and had a look along the first two meters with a microscope, it seems I have managed to clear the column of septum particles by removing 30 cm from the inlet end.
Sounds like the unknown student made an injection with a large bore needle, probably not even one made for an autosampler and the injection sent cores right into the inlet.
Either that or the solvent/sample that was injected contained massive amounts of siloxanes or dissolves siloxanes. If the sample does dissolve siloxanes it could have damaged the stationary phase itself.
Hi James,
I have caught up with the student this morning and it seems they may have used an unsuitable syringe to manually inject their samples, although this syringe has mysteriously disappeared without a trace so I cannot see for myself.
This looks pretty bad indeed. If you have the time for it, you can try to save the column by cutting 10-20 cm at a time. Worst case is a new column.
I would also perform a good inlet cleaning.
Hi Rndirk,
I have removed around 30 cm of column and that seems to have improved things somewhat:
Despite the imroved chromatogram it still doesn't look quite right, as the baseline is a bit all over the place and quite high. The run conditions of the instrument for that chromatogram are below:
INSTRUMENT CONTROL PARAMETERS: Nema1
---------------------------------------
C:\MassHunter\GCMS\1\methods\SPME Constant Flow.M
Thu Feb 23 11:48:42 2017
Control Information
------- -----------
Sample Inlet : GC
Injection Source : Manual
Mass Spectrometer : Enabled
No Sample Prep method has been assigned to this method.
GC
Oven
Temperature
Setpoint On
(Initial) 40 °C
Hold Time 2 min
Post Run 50 °C
Program
#1 Rate 15 °C/min
#1 Value 300 °C
#1 Hold Time 5 min
Equilibration Time 0.5 min
Max Temperature 325 °C
Maximum Temperature Override Disabled
Slow Fan Disabled
Cryo Off
ALS
Front Injector
Syringe Size 10 μL
Injection Volume 1 μL
Solvent A Washes (PreInj) 0
Solvent A Washes (PostInj) 0
Solvent A Volume 8 μL
Solvent B Washes (PreInj) 0
Solvent B Washes (PostInj) 0
Solvent B Volume 8 μL
Sample Washes 0
Sample Wash Volume 8 μL
Sample Pumps 6
Dwell Time (PreInj) 0 min
Dwell Time (PostInj) 0 min
Solvent Wash Draw Speed 300 μL/min
Solvent Wash Dispense Speed 3000 μL/min
Sample Wash Draw Speed 300 μL/min
Sample Wash Dispense Speed 3000 μL/min
Injection Dispense Speed 6000 μL/min
Viscosity Delay 0 sec
Sample Depth Disabled
Injection Type Standard
L1 Airgap 0.2 μL
Solvent Wash Mode A, B
Sample Overlap
Mode Sample overlap is not enabled
ALS Errors Pause for user interaction
Front SS Inlet He
Mode Splitless
Heater On 250 °C
Pressure On 22.123 psi
Total Flow On 104 mL/min
Septum Purge Flow On 3 mL/min
Gas Saver Off
Purge Flow to Split Vent 100 mL/min at 2 min
Liner Agilent 5190-2295: 870 μL (Universal, low pressure drop, ultra i)
Back SS Inlet He
Mode Splitless
Heater On 250 °C
Pressure On 16.555 psi
Total Flow On 18.2 mL/min
Septum Purge Flow On 3 mL/min
Gas Saver Off
Purge Flow to Split Vent 15 mL/min at 999.99 min
Thermal Aux 2 (MSD Transfer Line)
Temperature
Setpoint On
(Initial) 280 °C
Post Run 0 °C
Column
Column #1
Flow
Setpoint Off
(Initial) 1 mL/min
Post Run 0.78872 mL/min
Agilent 19091S-433: 1
HP-5ms
0 °C—325 °C (350 °C): 30 m x 250 μm x 0.25 μm
Column lock Unlocked
In Front SS Inlet He
Out PCM A
(Initial) 40 °C
Pressure 22.123 psi
Flow 1 mL/min
Average Velocity 15.761 cm/sec
Holdup Time 3.1723 min
Column #2
Flow
Setpoint Off
(Initial) 0.2 mL/min
Post Run 0 mL/min
Agilent 19091S-433: 2
HP-5ms
0 °C—325 °C (350 °C): 30 m x 250 μm x 0.25 μm
Column lock Unlocked
In Back SS Inlet He
Out PCM A
(Initial) 40 °C
Pressure 16.555 psi
Flow 0.2 mL/min
Average Velocity 3.4526 cm/sec
Holdup Time 14.482 min
Column #3
Pressure
Setpoint On
(Initial) 15 psi
Post Run 0 psi
Agilent
Retention Gap
0 °C—325 °C (325 °C): 2 m x 150 μm x 0 μm
Column lock Unlocked
In PCM A He
Out MSD
(Initial) 40 °C
Pressure 15 psi
Flow 3.6477 mL/min
Average Velocity 268.23 cm/sec
Holdup Time 0.012427 min
Column Outlet Pressure 0 psi
PCM A
PCM A He
PCM A He Supplies Column 3
Aux PCM A He
Pressure
Setpoint Off
(Initial) 10 psi
Post Run 0 psi
***Excluded from Affecting GC's Readiness State***
Signals
Signal #1: Test Plot
Description Test Plot
Details
Save Off
Data Rate 50 Hz
Dual Injection Assignment Front Sample
Signal #2: Test Plot
Description Test Plot
Details
Save Off
Data Rate 50 Hz
Dual Injection Assignment Back Sample
Signal #3: Test Plot
Description Test Plot
Details
Save Off
Data Rate 50 Hz
Dual Injection Assignment Back Sample
Signal #4: Test Plot
Description Test Plot
Details
Save Off
Data Rate 50 Hz
Dual Injection Assignment Back Sample
MS Information
-- -----------
General Information
------- -----------
Acquisistion Mode : Scan
Solvent Delay (minutes) : 0.00
Tune file : C:\MassHunter\GCMS\1\5977\ATUNE.U
EM Setting mode Delta : 0.000000
Normal or Fast Scanning : Normal Scanning
Trace Ion Detection : On
Run Time (if MS only) : 650 minutes
[Scan Parameters]
Start Time : 0.00
Low Mass : 50.00
High Mass : 350.00
Threshold : 50
A/D Samples: : 4
[MSZones]
MS Source : 230 C maximum 250 C
MS Quad : 150 C maximum 200 C
Timed Events
----- ------
Number Events= 0
END OF MS ACQUISTION PARAMETERS
TUNE PARAMETERS for SN: US1435L416
---------------------------------
Trace Ion Detection is ON.
EMISSION : 34.593
ENERGY : 70.007
REPELLER : 17.568
IONFOCUS : 90.331
ENTRANCE_LE : 25.226
EMVOLTS : 1135.323
Actual EMV : 1135.3
GAIN FACTOR : 1.58
AMUGAIN : 1504.000
AMUOFFSET : 123.750
FILAMENT : 1.000
DCPOLARITY : 1.000
ENTLENSOFFS : 12.901
MASSGAIN : -477.000
MASSOFFSET : -38.000
END OF TUNE PARAMETERS
----------------------
END OF INSTRUMENT CONTROL PARAMETERS
------------------------------------
Any further adive would be much appreciated, particularly with regards to correctly undertaking a thorough inlet cleaning.
Sorry for the very long post.
Best wishes,
Joe