Hi to everyone, I'm Massimo and I'm new to the forum.

I'm actually working in Italy (i'm italian, sorry for my english) in an environmental analysis lab.

In the lab we are working with a Gas Gromatograph Agilent, connetcet to a Teledyne Tekmar ATOMX for P&T analysis.

My problem is: When i run an analysis i have this weird column signal, and i find it in every sample, blank, or just water.
We have this signal until minute 5 and then it runs normally.

Any suggestions or advice?

Image
Hello

As you have GC coupled with Atomx sampler your problem has 2 possible sources: GC and/or Atomx. To narrow problem down I would disconnect GC from Atomx and check if your problem is not linked to GC column. But you probably have transferline combined with GC inlet so it will be tricky.
However you can check Atomx - make sure that your analytical trap is baked out (or put new one, it could be saturated and the only way is to replace it).

Try to run method with no injection from Atomx (just GC method) - at least it will show you if GC column needs baking out.

Regards

Tomasz Kubowicz
2 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry