Latanoprost
Posted: Mon Feb 13, 2017 10:07 am
by tinsang
Dear all,
I analyze a drug which contain the latanaprost molecule using GC and have a great difficulty to identify this molecule because I did not find that topic in literature.
I used column DB-1701,30mx0,25mmx0,25 µm.
Oven temp. 80°C, hold 1 min; 30°C/min, 160°C, 10°C/min, 280°C, hold 20min.
Inj: 280°C
Det.: 280°C
Flow : 1ml/min He
Split 1/10
Solvent: 2- propanol
Ref. solution: 1 mg Latanoprost/1ml 2- propanol
Latanoprost showed up for the first time at 17 min (5 Injections). But for second sequence Latanoprost disappeared.
Was it pure coincidence? Any suggestions?
Re: Latanoprost
Posted: Mon Feb 13, 2017 8:36 pm
by Rndirk
If you want to analyze anything using GC, one rather fundamental requirement is that the molecule is present as a gas at the temperatures you're using.
Have you considered this for your analysis?
Also, what kind of detection do you use?
Re: Latanoprost
Posted: Mon Feb 13, 2017 10:55 pm
by James_Ball
Looked it up here http://www.lookchem.com/Latanoprost/
Looks like a boiling point of 583.8C, unless it has a high vapor pressure below the boiling point it is probably going to be difficult to do by GC.
Re: Latanoprost
Posted: Tue Feb 14, 2017 11:42 am
by tinsang
Thanks for the hints. I have mixed up the flash point and the boiling point. :cry:
Re: Latanoprost
Posted: Tue Feb 14, 2017 12:03 pm
by Rndirk
It looks much better suited for liquid chromatography :wink: