internal standard in LC MS MS
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
5 posts
Page 1 of 1
Hi! Is it possible to use an internal standard in ESI negative when me analyte is in ESI positive instrument method?
Magda
-
- Posts: 1074
- Joined: Mon Aug 30, 2004 5:42 pm
I do not quite understand your question (maybe you could clarify) but theoretically you could use an internal standard that produce molecular ions in both positive and negative ionization modes (i.e. zwitterionic compounds is a good example...).
-
- Posts: 5
- Joined: Fri May 19, 2006 5:55 am
I mean, Is it valid to have a quantification of a substanse which is been measured in ESI+ by using an internal standard measured in ESI-, (they are well seperated in my current Chromatography conditions). This is because I want to avoid the addition of another compound in my analysis (I have already 3: two analytes ESI+ and ESI- and one internal ESI-)
So In other words is it possible to use the existing internal for the quantitation of both analytes?[quote][/quote]
So In other words is it possible to use the existing internal for the quantitation of both analytes?[quote][/quote]
Magda
-
- Posts: 1074
- Joined: Mon Aug 30, 2004 5:42 pm
I see now what you mean...
I would say that in this case you would be able to correct for general drifting of the instrument, however, ion suppression effects might be totally different in positive vs. negative mode and your internal standard won't be able to correct for those...
I would say that in this case you would be able to correct for general drifting of the instrument, however, ion suppression effects might be totally different in positive vs. negative mode and your internal standard won't be able to correct for those...
-
- Posts: 974
- Joined: Tue Jan 11, 2005 11:37 pm
Your scheme will correct for variabilities between the time of addition of the internal standard and the time of the sample injection. After that, the assumptions of internal standard calibration may no longer hold. The big assumption is that the IS and the analyte behave the same way in the detector. Until you prove that, your method is not validated. In general, you can't assume that the MS source is well behaved, but you can prove it.
Mark Tracy
Senior Chemist
Dionex Corp.
Senior Chemist
Dionex Corp.
5 posts
Page 1 of 1
Who is online
In total there are 220 users online :: 0 registered, 0 hidden and 220 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: No registered users and 220 guests
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: No registered users and 220 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science