Problems with TCD analysis of trace nitrogen in hydrogen
Posted: Fri May 19, 2006 3:41 pm
First the settings:
HP5890 w/ TCD, helium carrier, high TCD sensitivity, [+] signal setting, 30C isothermal run, Valco Sampling valve, tried sample loops from 250ul to 5mL, tried both a 1.5m Molsieve 13x and a 30' Hayesep DB
I run a 100ppm Nitrogen standard and have only my single nitrogen peak. If I run 'UHP' Hydrogen, however, I get my hydrogen coming off as a negative peak, but then at the time nitrogen should elute I get another, small negative peak instead of a positive peak one would expect from nitrogen. As mentioned above, I have tried two different columns to rule out any kind of flow upset coeluting with nitrogen, but the negative peak tracks with nitrogen on both columns. I have tried injections at both atmospheric and at carrier pressure (20psi for the molsieve), as well as sample loops ranging from 250ul to 5mL in size, and the negative peak size tracks with the amount of sample being injected. What can be causing this negative peak? If I had nitrogen as an impurity in my helium carrier, could conditions exist that give me a positive peak for nitrogen at 100ppm, but negative at a lower concentration, for example 10-20ppm?
HP5890 w/ TCD, helium carrier, high TCD sensitivity, [+] signal setting, 30C isothermal run, Valco Sampling valve, tried sample loops from 250ul to 5mL, tried both a 1.5m Molsieve 13x and a 30' Hayesep DB
I run a 100ppm Nitrogen standard and have only my single nitrogen peak. If I run 'UHP' Hydrogen, however, I get my hydrogen coming off as a negative peak, but then at the time nitrogen should elute I get another, small negative peak instead of a positive peak one would expect from nitrogen. As mentioned above, I have tried two different columns to rule out any kind of flow upset coeluting with nitrogen, but the negative peak tracks with nitrogen on both columns. I have tried injections at both atmospheric and at carrier pressure (20psi for the molsieve), as well as sample loops ranging from 250ul to 5mL in size, and the negative peak size tracks with the amount of sample being injected. What can be causing this negative peak? If I had nitrogen as an impurity in my helium carrier, could conditions exist that give me a positive peak for nitrogen at 100ppm, but negative at a lower concentration, for example 10-20ppm?