By Jaime on Wednesday, August 11, 2004 - 07:43 am:
When determining the precision of a method, do you make 4 or more sample preparation and one injection each, or one sample preparation and 4 or more injection from that preparation. Or is it 4 sample preparation or more and 4 0r more injection per sample preparation.
By David Blais on Wednesday, August 11, 2004 - 08:52 am:
Jaime,
Method precision should be multiple sample preps and can be one injection each. Whenever you perform multiple injections from one preparation, you are looking at autosampler precision. Typically when validating a method I prepare six sample preparations and make one injection of each.
Hope this helps.
-David
By Jaime on Wednesday, August 11, 2004 - 11:26 am:
Thank you David.
By Leadna N Brigola on Friday, August 13, 2004 - 12:49 am:
Jaime,
Agree with David.Method precision is done using at least 6 replicate preparations of the sample and single injection of each replicate. for system precision its 6injections of 1 solution.
Hope this helps.