camino » Mon May 08, 2006 10:31 am
hello,
You should add the IS after the extracion proccess, directly to the pentane fase after the extraction.
In the L-L extraction there is a problem to known the final volumen of the organic phase. i.e If you use 100 mL of sample and 10 mL of hexane, after the extraction you don't obtain 10 mL of hexane, you get about 9 mL, due you lose solvent because it dissolve with the water and/or it evaporate; and when you calculate the original concentration in the water, how we can calculate the pre-concentratio factor???
If you don't use IS the recovery will be higher (~ 120 - 140%).
This only can resolve working with IS, or extracting the calibration standards, trating it as samples. But this second option is more time consuming. Normally we work with external calibration (the standard are prepared in the same solvent as extraction, rising at determinated volume, and adding IS).
I add 200 µL of IS (5 mg/L) to the hexane directly, in this way the IS correct the loss of hexane (or pentane) and the recovery take down to normal values (~80-90%). Then I take 1 mL to a chromatographic vial, and inject 1 µL in split mode. I use 35ms capillary column.
You may add the IS before the extraction, directly to the sample (the IS must be in a water miscible solvent i.e methanol), and treat it as surrogate.
I hope you understand all, I'm practising my English actually. Tell my if you need any explanation.
Best regard
Best regard