Discrimination factor
Posted: Mon May 15, 2006 5:30 pm
by LY
Who Knows What Is Discrimination factor between the peaks ?
Posted: Mon May 15, 2006 9:59 pm
by tom jupille
To find out, use the Search function here on the Forum. Enter "discrimination factor" as the search term, and make sure you have "search for all terms" selected.
Posted: Tue May 16, 2006 2:15 pm
by LY
Thanks, Tom,
I did it yesterday before posting my question on this site, but search result was: "No topics or posts met your search criteria".
Posted: Tue May 16, 2006 4:05 pm
by Rob Burgess
I can confer with Tom that there is a "thread" related to this question if you search all terms...
Posted: Tue May 16, 2006 5:03 pm
by LY
OK, you are right, thanks.
“… discrimination factor where you divide the peak height - the height of the valley by the total peak height..
So, according to this explanation, Discrimination factor = Peak-to-valley ratio?
Posted: Tue May 16, 2006 10:22 pm
by Uwe Neue
The specific definition of the discrimination factor (from my book) is the ratio of the distance between the apex of the smaller peak and the valley to the height of the smaller peak.
Further discussion in Uwe Neue, HPLC Columns, pages 303 to 308, or Fallah and Martin, Chromatographia 24 (1987), 115.
Posted: Tue May 16, 2006 11:03 pm
by tom jupille
LY, I apologize for my tone.
I had just come across several other posts which had already been discussed, and it had been a long day (at least, that's my excuse!).
Posted: Wed May 17, 2006 4:07 pm
by LY
That’s absolutely fine, Tom, and thanks again for you consideration.