Doesn't your instrument software have an inlet setting called "solvent vent" or "large volume injection (LVI)"?
It does exactly what Peter and James pointed out: split during and shortly after injection (evaporation of solvent), then splitless heating to get everything on column, then split again once everything is on the column.
Like i said in my first post, the parameters of this process should be carefully chosen/optimized or you risk backflash or losing analytes.
A "huge drop" in vacuum => this tells me that too much solvent is injected. So depending on your liner volume, you'll also have backflash.
A drop in vacuum can be expected, I even see it when doing a 1ul splitless injection with DCM, but also when doing cold splitless or solvent vent since part of the process is to transfer a little solvent to a cold retention gap to get a proper solvent effect for the lighter compounds. You hold the filaments off until after the solvent front passes through the system. If timed correctly you will be back to normal operating vacuum just before the first analytes elute.
The fact that the interference occurs only with extracted samples or spiked extracted samples and not standards spiked into pure solvent says there is something in the extraction process that is causing the problem here. Either excess moisture or a mixture of solvents, or an extractable interference is present. Hopefully the full scan run will help to show what the interference could be.