Does ammonium carbonate pH 10 destroy pumps?
Posted: Wed Feb 01, 2017 1:10 pm
Dear all
We are struggling with dying pumps here at the moment. It all started with an Agilent 1100, mixing 75 % ACN with 25 % ammonium carbonate (50 mM pH 10). This was fine for one hour or two, then we saw big pressure ripples (> 10 bar) soon followed by complete failure: The pistons moved, but no liquid was pumped at all.
I changed the inlet valve because I thought the liquid would flow back instead of forward due to a leak in there => Did not help.
I exchanged the pump seals (they were 2 months old, but anyway) => Pumping worked again, but again only for some hours.
Then we had a service technician here who exchanged the whole pump head => was fine for some hours, same problem again
Then we had the same technician, exchanging the whole pump to a binary model => was fine for some hours, then same problem again
I then transferred the method to an older model (Gynkotek M480). Same thing there. Pressure is stable for about two hours, then failure. At least this pump works fine with methanol/water after purging. The Agilents are all dead now.
So my question of the year: What is it in ammonium carbonate that kills the Agilent pumps? And how can I revive them?
Kind regards
Jörg
Kind regards
Jörg