We have an LC system coupled to a QTOF mass spectrometer. Few days ago the column got jammed and the pressure started building beyond the max pressure limit. I tried washing the column with 95% and 100% acetonitrile... after that it seemed okay.. but even then the pressure was higher than usual.
Last week I was washing the column with 95% ACN, suddenly an error showed... "Compressor limit of the pump exceeded". I Googled it and got the suggestion to purge the pump. After purging, when I closed the purge valve, I got no pressure. Certainly, somewhere it is leaking. Then I saw a point on the sample loop is bulged up and solvent is leaking from there. I have attached the picture for your reference.
My question is should I just change the sample loop and the column? Or there is a serious need for a more in depth inspection? The company people are asking for huge amount of money for inspection. We don't have an AMC also.
I am also curious about what made this loop rupture happen, so that I can avoid it in future.
If I just go for a sample loop what should be the pore size of the loop? And also, We run various types of sample, from peptides to small molecules. What pore size column should I go for such a general purpose use?
Thanks in advance
Arnab