6890/5973 GC/MS quad temperature fault

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hello,

I am using an HP 6890/5973 GC/MS. The quadrapole temperature at times will jump up very quickly to 511 C and then shut off giving a quadrapole temperature fault. This seems to happen especially as the oven ramps, it will stay steady at 150 up until the oven reaches a certain temp and then it will start shooting up. I have checked the sensor to no avail. Has anybody ever heard of something like this?

Thanks!
I had that happen on my 5975. It would read correctly 99% of the time, the other 1% it would wander all over the place. A new sensor and heater fixed it.
Hello,

The quadrapole temperature at times will jump up very quickly to 511 C and then shut off
The 511C is almost certainly a problem with the sensor (open circuit)
The instrument I use reads 512C when temperature sensors are open circuit.
512 is 2E9 which is probably not a coincidence!
3 posts Page 1 of 1

Who is online

In total there are 25 users online :: 0 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry