Page 1 of 1

SPD-M10Avp PDA detector tungsten lamp

Posted: Wed Jan 25, 2017 1:32 pm
by John Jones
Hi,

we have a number of HPLC systems in the lab here.

In particular i have a PDA detector Code SPD-M10Avp

My question is what is the Tungsten lamp specifications for this system?

What is the voltage and wattage required in this bulb?

It does not say on the lamp taken out of the system, and is not in the manual.

Thank you very much.

Regards

John

Re: SPD-M10Avp PDA detector tungsten lamp

Posted: Wed Jan 25, 2017 5:41 pm
by uzman
It is a 12 volts / 20 Watt tungsten lamp

Re: SPD-M10Avp PDA detector tungsten lamp

Posted: Tue Feb 07, 2017 6:37 pm
by John Jones
Thank You

Regards.

John