Reaction/color change of Diethylamine
Posted: Tue Jan 24, 2017 7:54 pm
by adam
Hello
In our lab, we have seen it happen a few times when diethylamine is used as an additive in the mobile phases, that the mobile phase will undergo a color change over time.
Can anyone shed any light on this? Can anyone explain exactly what is the reaction that's causing this?
We are trying to understand the mechanism, so that we can perhaps avoid using DEA in combination with things it doesn't "get along with".
Much Thanks in Advance
Re: Reaction/color change of Diethylamine
Posted: Wed Jan 25, 2017 4:17 pm
by GOM
Hi Adam
What is the colour change?
What is the mobile phase in which this occurs?
If a mixed phase, have you tried the diethylamine with each component separately to see which one is giving the colour change?
Regards
Ralph
Posted: Thu Jan 26, 2017 1:24 am
by adam
Thanks GOM
We have seen this on two occasions.
- In the first case it was 1% DEA in acetonitrile, with a small concentration of naphthol as internal standard.
- In the second case it was 1% DEA in 90/10 acetonitrile/ethyl acetate
I tend to think it was not reacting with these ingredients so much as with a trace impurity......or itself.......of oxygen.
Any thoughts, or any other experiences that folks may want to share, would be appreciated.
Thanks!!
by adam
Oh, and in response to your other question, the color changed from clear to yellow.
Posted: Thu Jan 26, 2017 3:28 am
by HPLC chemist
The DEA is reacting with the Ethyl Acetate to give you an amide.
Posted: Fri Jan 27, 2017 1:14 am
by adam
Thank you for that reply, but how do you account for the color change in the other solution (just acetonitrile and a low concentration of naphthol). Do you think it might have reacted with the naphthol?
Am I the only one who has ever noticed this kind of color change with Diethylamine?
Posted: Fri Jan 27, 2017 7:41 am
by tkubowicz
Hello
Check if glasware (bottles) is really clean, especialy if you're using washing machine - perhaps there are traces of some detergents or cleaning agents.
Regards
Tomasz Kubowicz
Posted: Fri Jan 27, 2017 8:04 am
by Forse
DEA will get a yellow tint on its own given enough time.
At my last workplace we had bottles from the 70's and 80's which had turned different degrees of yellow.
