peak areas are half size than normally

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hello Everybody.

We are running EP rel substances and something strange happened.

The peak areas is almost half the size than normally. I asked another analyst to prepare a new standard , but the peak size is still half of the normal size. There is no shift in the retention times, could not see any leaks either. The injection volume was set up to correct one.
All other peak , the solution peak on the start of the chromatogram is also half of the size...
any ideas ? I don't think that this is due incorrectly prepared mobile phase either..
Hello

Put some information about instrument (brand, model etc)
It could be "injection system" depends what you have. I'd check:

-injection valve (rotor seal)
-syringe (if you have one)
-sample loop
-metering device

Regards

Tomasz Kubowicz
Before checking ^ - have a look at the instrument method - any recent changes to sampling rate?
Thanks,
DR
Image
3 posts Page 1 of 1

Who is online

In total there are 17 users online :: 2 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry